, and I don't understand why the surface reflectance (rhos) at 709 nm is always depleted as compared to the two nearby wavelengths (681 and 754). I speak about ocean spectrum, far from sunglint or clouds, with no vicarious gain applied. See enclosed png file. I expected a 709 nm rhos value approximately given by a linear interpolation between 681 and 754 rhos values. Is there some correction missing in Seadas for this particular wavelength ?

Thanks

Jacques

Re: OLCI 709 nm band

Posted: Tue Oct 05, 2021 4:15 pm America/New_York
by OB.DAAC - SeanBailey
Vicarious calibration....or rather the lack thereof. Once properly calibrated, the result is more like what you would expect:
Screen Shot 2021-10-05 at 4.11.34 PM.png
Screen Shot 2021-10-05 at 4.11.34 PM.png (82.39 KiB) Not viewed yet
The above was generated from a scene using the following gains:
gain=[0.9795,0.9953,0.9796,0.9675,0.9697,0.9695,0.9750,0.9750,0.9744,0.9788,1.0060,0.9834,1.0,1.0,1.0755,0.9908,1.0,1.0174,1.0,1.0,1.0]

Keep in mind these are not likely to be the set of gains we apply when we finally get around to distributing L2 data...but they're probably close.

Sean

Re: OLCI 709 nm band

Posted: Wed Oct 06, 2021 8:20 am America/New_York
by jstum
Thanks Sean. Is there some internal report about the method used to derive these coefficients ?

Re: OLCI 709 nm band

Posted: Wed Oct 06, 2021 9:01 am America/New_York
by jstum

Re: OLCI 709 nm band

Posted: Wed Oct 06, 2021 10:00 am America/New_York
by OB.DAAC - SeanBailey
Yes, that would be the answer :D